The scan rate of a spectrum analyzer or a scanning measuring set should be adjusted for the
CISPR frequency band and the detection mode used.
The minimum sweep time/frequency or the fastest scan rate is listed in the following table: 最小扫描时间/频率即最快扫描速率
Frequency band
Peak-detection
Quasi-peak detection
A
9 kHz – 150 kHz
100 ms/kHz
20 s/kHz
B
0,15 MHz – 30 MHz
100 ms/MHz
200 s/MHz
C/D
30 MHz – 1 000 MHz
1 ms/MHz
20 s/MHz
For a spectrum analyzer or scanning measuring set used in a fixed tuned non-scanning mode,
the display sweep time may be adjusted independently of the detection mode and according to
the needs for observing the behaviour of the emission. If the level of disturbance is not steady,
the reading on the measuring set must be observed for at least 15 s to determine the maximum.
Minimum scan times for the three CISPR bands with peak and quasi-peak detectors 使用峰值和准峰值检波器时的最小扫频时间Ts